TomoEQA: Dose verification for patient-specific quality assurance in helical tomotherapy using an exit detector
Journal
Physica Medica: European Journal of Medical Physics
Vol
82
Author
Min Cheol Han, Kyung Hwan Chang, Jihun Kim, Su Chul Han, Kwangwoo Park, Dong Wook Kim, Hojin Kim, Jin Sung Kim
Year
Miscellaneous
Date
Jan. 2021
Purpose

Existing phantom-less quality assurance (QA) platforms does not provide patient-specific QA for helical tomotherapy (HT). A new system, called TomoEQA, is presented to facilitate this using the leaf open time (LOT) of a binary multi-leaf collimator, as measured by an exit detector.
Methods

TomoEQA was designed to provide measurement-based LOTs based on detector data and to generate a new digital imaging and communication in medicine (DICOM) dataset that includes the measured LOTs for use by secondary check platforms. To evaluate the system, 20 patient-specific QAs were performed using the program in Mobius3D software, and the results were compared to conventional phantom-based QA results.
Results

From our assessment, most of the differences between the planned and measured (or calculated) data, excluding one case, were within the acceptance criteria comparing with those of conventional QA. Regarding the gamma analysis, all results considered in this study were within the acceptance criteria. In addition, the developed system was performed for a failed case and showed approximately the same trends as the conventional approach.
Conclusions

TomoEQA could perform patient-specific QAs of HT using Mobius3D and provide reliable patient-specific QAs results by evaluating point dose errors and 3D gamma passing rates. TomoEQA could also distinguish whether an intensity-modulated radiation therapy plan failed or not.
Keywords

Helical tomotherapy, Patient-specific quality assurance, Phantom-less quality assurance, Leaf open time, Mobius3D